AccuCT with this setup? #141
-
Hello! I am looking to switch away from IoTaWatt with their recent announcement and have a working setup with AccuCT clamps. Was trying to save money and reuse these clamps if possible. Here is their data sheet - https://www.accuenergy.com/wp-content/uploads/acuct-hinged-series-compact-split-core-current-transformer-datasheet.pdf I have the 50A (12), 100A (2), and 200A (2) models currently in use today. I am in the US so the two 200As are on my Main feed coming into the power company and then I have the 50As spread across various circuits. I assume, based on the FAQ, and the data sheet these require me to severe the jumper since they have burden resistors right? Anyone else may have calibrated these units? Just trying to make the switch as seamless as possible on some known circuits. |
Beta Was this translation helpful? Give feedback.
Replies: 2 comments 6 replies
-
I didn't realize that IoTaWatt was not being made anymore. Thanks for letting me know. It's a shame, because from what I gathered that was a great product. For the CTs, it looks like the AccuCT's all output 333mV when they are reading their rated peak. Since they have an internal burden resistor, you'll just need to sever the jumpers on the back of the board to disable the on-board burden resistors. You'll also need to calibrate the CTs to the meter using the method here: https://www.youtube.com/watch?v=BOgy6QbfeZk&t=1261s You can start with a value of 5132 for the 50A CTs. |
Beta Was this translation helpful? Give feedback.
-
As a test with my AccuCT’s I tried severing the JP1 pad before doing any testing and it didn’t work. I even ordered a bunch of the YHDC’s from AliExpress and they didn’t work in ct1 till soldered the jp1 back.
Test before you sever first, that was my experience. Also make sure you have a know good and repeatable load testing to calibrate as needed.
-
Richie
On Jun 18, 2023, at 1:34 PM, ICPete ***@***.***> wrote:
Hi John,
Since I'm also interested in using alternative CTs, I'm trying to understand what will work and what won't.
I understand your circuit uses onboard 22 ohm burden resistors, which can be removed by severing their jumpers.
With the onboard resistor in place, we need a CT outputting a maximum current of 33 mA, correct?
With the onboard resistor severed, we need a CT outputting a maximum voltage of 720 mV, correct?
For reference, I notice from the AcuCT Hinged Series datasheet that their built-in burden resistors are as follows:
50A: 20R
100A: 10R
200A: 5R
Since they are all designed to output a maximum voltage of 333 mV, I conclude that their secondary max currents are as follows:
50A: 16.65 mA
100A: 33.3 mA
200A: 66.6 mA
To me that implies they all have a 3000:1 turns ratio. Just doing the calculation for fun.
Finally, if I were to locate a 200A CT that output, say, 100 mA max (i.e., 2000:1 turns ratio, which is quite common on AliExpress), could I place a resistor in parallel with the onboard 22 ohm resistor to bring the combined resistance down to, say, 7.2 ohms, in order to produce a maximum voltage of 720 mV? By my calculation that would require the added resistor to be 10.7 ohms.
John, thank you for bringing this excellent circuit to fruition, and for making it available to us!
—
Reply to this email directly, view it on GitHub<#141 (reply in thread)>, or unsubscribe<https://github.com/notifications/unsubscribe-auth/AAA2RH5A4IXTMO7XXEXFEJDXL43Y5ANCNFSM6AAAAAAY3D4OYU>.
You are receiving this because you authored the thread.Message ID: ***@***.***>
|
Beta Was this translation helpful? Give feedback.
I didn't realize that IoTaWatt was not being made anymore. Thanks for letting me know. It's a shame, because from what I gathered that was a great product.
For the CTs, it looks like the AccuCT's all output 333mV when they are reading their rated peak. Since they have an internal burden resistor, you'll just need to sever the jumpers on the back of the board to disable the on-board burden resistors.
You'll also need to calibrate the CTs to the meter using the method here: https://www.youtube.com/watch?v=BOgy6QbfeZk&t=1261s
You can start with a value of 5132 for the 50A CTs.